External ADC input unresponsive after 12V high-side measurement
I am looking for advice regarding an issue with the external ADC on my Otii Arc.
During a recent IoT hardware test, I needed to perform a differential current measurement using an external 100mOhm shunt resistor. I’m following this guide Differential measurements - Use cases. but Unfortunately, without realizing the 5.5V maximum rating on the ADC pins at the time, I placed the shunt resistor on the high-side of a 12V battery supply.
I actually managed to perform this 12V high-side measurement 27 times. During those tests, the current was somehow registering, but the voltage was only reading 1.3V (which I now realize was completely incorrect given the 12V supply). since I can’t attach otii3 file I am attaching one of the results from 27 measurements on screenshot below.
Currently, the external ADC seems to have stopped functioning properly. I have since re-tested the setup using 4.2V battery but the ADC current is no longer reading the actual load. Instead, as seen in the attached screenshot, the ADC current reading just shows a flat, noisy signal ranging from approximately -175 µA to 75 µA, and the voltage is reading around 1.59 mV.
I suspect that the repeated 12V exposure has damaged the internal circuitry on the expansion port.
Could you please advise on what my best options are for this unit? Is there an official repair service, or a specific diagnostic procedure you recommend I follow to confirm the extent of the damage?
Hi and welcome to the forum.
Yes, I agree with your analysis, I believe that the ADC input protection is burned.
We have a repair and calibration service (350USD + freight for Arc) and the unit needs then to be shipped to us in Sweden.
If you are skilled in soldering, you can try to repair it yourself, with instructions from us.
I recommend that you create a case in our case system ( User Management - Qoitech ) and we will handle it there.